Maximum Operating Temperature of the Analog Discovery 2


Commanderfranz

Question

1 answer to this question

Recommended Posts

Hi @Commanderfranz

I can't find any specification for maximum temperature for AD2. I suppose we can consider the common 0...40*C standard.

The devices are preheated for manufacturing test and calibration procedure, which is performed at normal room temperature (25*C), having the internal device temperature around 45*C
The temperature drift of the scope reading is about (40 ppm/*C).
In case you want better precision the device can be recalibrated using the WF application. Before this, leave the device connected with the application for 10-30 minutes to reach stable operating temperature.