LOGGER Values are incorrect but the measurements using Scope is correct


naveen.george

Question

4 answers to this question

Recommended Posts

Hi @naveen.george

The Logger and Voltmeter by default use the highest range, 50V pk2pk on AD, with resolution of ~3mV.
In Scope you set the range to 10mV/div, which uses the 5V pk2pk range on AD, with 300uV resolution.
Beside the resolution in high range (50Vpk2pk) the calibration error and temperature drift is higher.

Link to comment