Measuring a 110 VDC signal on Analogue Discovery 2


AndyHook1970

Question

Just two quick questions from a newbie.

1) Will I be able to measure on the Scope a 100VDC signal that is derived from mains?  I plan to use a 10:1 probe.  

2) Can I cause any damage by exceeding the 25V input limit on the analogue inputs?

 

Many Thanks

Andy

Link to comment
Share on other sites

1 answer to this question

Recommended Posts

Hi @AndyHook1970

2. The recommended maximum voltage for scope inputs of Analog Discovery (1&2) is ±25V (50V differential) and the absolute maximum is ±50V.
Exceeding the abs max could damage the device. See:
https://reference.digilentinc.com/reference/instrumentation/analog-discovery-2/reference-manual#scope_input_divider_and_gain_selection

1. With 10x probe you could measure up to ±250V, but be cautions when working with high voltages. 

To see the real value, under channel options configure the applied attenuation:
i1.png.9e9c2e19332a86b6070438905bf1df3e.png