Orycho Posted December 7, 2019 Share Posted December 7, 2019 I connect wavegen output pin to the oscolloscope pin directly. The square wave distort at 1MHz but the sine wave is normal. Is this normal or I have to calibrate it. Link to comment Share on other sites More sharing options...
attila Posted December 9, 2019 Share Posted December 9, 2019 Hi @Orycho This is called ringing: https://en.wikipedia.org/wiki/Ringing_(signal) To reduce this you should use use a resistor, on the generator side. To remove it use BNC adapter, cabling and 50 ohm termination. Link to comment Share on other sites More sharing options...
Question
Orycho
I connect wavegen output pin to the oscolloscope pin directly. The square wave distort at 1MHz but the sine wave is normal. Is this normal or I have to calibrate it.