Analog Discovery 2 vs new Raspberry Pi 4 (4GB)


FloMai

Question

Hey guys it was announced that with the new RPi 4 the connection problems with AD2 where solved, but it still seems not stable ERC: 0x2.

Any suggestion how to improve setup?

Power supply is 5V/3A and AS2 is connected to USB 3.0 port.

Linux version 4.19.75-v7l+ / Raspbian GNU/Linux 10 \n \l


2019-11-26-150916_1920x1080_scrot.thumb.png.caebb221f81a7e77e32f92b8f21144f2.png

BR  

 

Link to comment
Share on other sites

10 answers to this question

Recommended Posts

Hi @FloMai,

I apologize for the delay. I updated my Raspbian desktop to the same edition (4.19.75) but I am able to detect and run the Analog Discovery 2 with a Raspberry Pi 4 Model B without. Could you attach a picture of your setup? Otherwise, I might recommend trying an external power supply. @attila do you have another idea of something this user might try?

Thanks,
JColvin

Link to comment
Share on other sites

Hi,

I have done lots of testing, after >one day triggering logic signals with low sample rate connection fails. Also when I just trigger one channel, so I guess its not a thing of to many data.

What I already tried is:

- different external power supplies (1-3Amps)

- change from USB3.0 to USB2.0 port ( in the beginning it seemed that's stable but...)

- operation without keyboard and mouse load

For me it looks like there is still some problem in HW, may USB of RPI is not fully USB compliance, I will go on with testing...

setup.thumb.jpg.6f126b2a275da0c5c1b858081ea55760.jpgdesktop.thumb.jpeg.159f15f4f10dee9416fec06eba6c1d34.jpeg

 

could you do some long term test?

 

BR Florian

 

Link to comment
Share on other sites

Hi @FloMai

I have tested RPi4 for about week.
Running continuously for 1-2 days combinations of USB 3 and 2.0 ports, with hub and without it.

The only issue I noticed was with hub connected to the USB 2.0 port. This could have been an issue with the hub. I have had problems in the past with specific hubs even when connected to the PC, random disconnects with some devices...

 

Link to comment
Share on other sites

Hi Attila,

thanks for the reply, I have tested now with USB 3.0 hub (Anker 7-Port) and connection failes anyway after some hours, I doublechecked with 2nd setup (RPI4 and AD2) and same failure.

how is your setup in waveform, did you run a trigger over days or just run scope?

can you open logic register, add all 16 channels and set trigger for all in long term?

 

BR Florian

Link to comment
Share on other sites

Hi @FloMai

Sorry for the late reply. It looks like event the RP4 has problems with the AD2. 
In my earlier tests where it was running for 1-2 days, the communication rate with the device was only about 1/second.
Repeating the test with ~100/second rate, I get 0x02/timeout error after about a day with USB 2 and 3 ports, and with USB hub every 6-12 hours.

Link to comment
Share on other sites

HI @attila

ok thanks, is it possible to program a connection check or reconnection loop via script?

Or has anybody programmed some labview application with no connection issue?

Do you think with 1/second it will run over long term, I don't need a high sampling?

Link to comment
Share on other sites

Hi @FloMai

The issue is that the USB packets get lots, the communication gets out of sync.
This should also happen with custom application/script, but with this it is possible to reconnect to the device.
I will try to add to some retry or reconnect mechanism to the library.
With lower communication rate, ~1 per second it was running for me 1-2 days without any issue.