I'm using WaveForms 3.8.22's Spectrum Analyzer with Analog Discovery 2.
When I run the analyzer I use FFT with 129 BINs in the 1 Hz to 100 Hz range on both channels.
The magnitude settings of the Spectrum Analyzer are the follows: Vrms, top: 10 mV, bottom: 0 V, reference: 1V, range: 5V.
I selected the exponential dB Average on both channels.
If I run the spectrum analyzer and while it's running I change the top value from 10 mV to let's say 20 mV it instantly changes the chart correctly, but after a new measurement completes the chart changes drastically which is not the expected behavior, I think.
Here is a screenshot of my workspace (just before the change):
And another one just after the change from 10 mV to 20 mV:
As you can see the chart changes significantly while the source signal remained somewhat the same.
Interestingly when I tried to reproduce the problem by connecting the signal generator with the scope while generating a steady sine signal, the problem didn't show.
It did show though when I set the signal generator to sweep mode.
It also work fine if I change the type of the trace from the exponential dB average to sample.
Is there a reason why this could happen? Does the averaging algorithm's buffer get reset when we change any of the parameters of the spectrum analyzer?
Question
Andras
Hi,
I'm using WaveForms 3.8.22's Spectrum Analyzer with Analog Discovery 2.
When I run the analyzer I use FFT with 129 BINs in the 1 Hz to 100 Hz range on both channels.
The magnitude settings of the Spectrum Analyzer are the follows: Vrms, top: 10 mV, bottom: 0 V, reference: 1V, range: 5V.
I selected the exponential dB Average on both channels.
If I run the spectrum analyzer and while it's running I change the top value from 10 mV to let's say 20 mV it instantly changes the chart correctly, but after a new measurement completes the chart changes drastically which is not the expected behavior, I think.
Here is a screenshot of my workspace (just before the change):
And another one just after the change from 10 mV to 20 mV:
As you can see the chart changes significantly while the source signal remained somewhat the same.
Interestingly when I tried to reproduce the problem by connecting the signal generator with the scope while generating a steady sine signal, the problem didn't show.
It did show though when I set the signal generator to sweep mode.
It also work fine if I change the type of the trace from the exponential dB average to sample.
Is there a reason why this could happen? Does the averaging algorithm's buffer get reset when we change any of the parameters of the spectrum analyzer?
Thank you,
Andras