AD2 Trigger out pulse length


msiklodi

Question

Dear colleagues,

is there any possibility to set(configure) the trigger output pulse length? I have two discoverys, and one somehow provides 125us, while the other 350us. They might have been set for different value, but I do not know ho to (re)set them. 

 Thanks for your help in advance,

Miklos

Link to comment
Share on other sites

4 answers to this question

Recommended Posts

43 minutes ago, attila said:

Szia @msiklodi

The trigger output length depends on the selected source and it is not adjustable.

In this setup Trigger 1 outputs the Scope Detector, which is wired to DIO 0, see the 10ns pulse.
You can use the Pattern Generator to output a configured pulse length, like DIO 1 is set to 0.3us

image.png.5f855dbcfb741796ff75d13b8d995d90.png

 

Link to comment
Share on other sites

Aha the Device Manager has to be set to a mode which supports "patterns". I had the AD2 set to the only option with 16K scope sample depth, since that is important to us...

I would like to trigger a 2nd AD2 module when the first one triggers, and also trigger an external device (Dataq 718) when the first AD2 triggers.