I encountered a peculiar issue when using AWG2 of the Analog Discovery 2.

Physical connections were as follows:

 1+ to W2

 1- to GND

When AWG2 is set to generate a sine wave voltage <= 1.00V, the output seems to be a clipped version of a voltage referenced to +5V (USB +ve rail).

At any voltage above that (>1.001), the output is fine. 

I'm using Waveforms 3.8.2 on a 64bit Win10 PC.

I've  attached a screen capture:

2018-07-19_16-33-58.thumb.gif.58aec0724aec7eae29598ac7e0312a1f.gif

Link to comment
Share on other sites

2 answers to this question

Recommended Posts

Hi @avnrdf

I suppose the AWG2 is wrongly calibrated at low gain. Wavegen 2 LG additive parameter should be a few millivolts (like 0.005, 0.016) but on your device is around 5.5

Open WaveForms/ Settings/ Device Manager/ Calibration and select Reset/Load factory

image.png.351bd32d4cf9140a745b1dbdf71117fc.png