problem with triggering low frequency signals

Question

I have trouble correctly triggering signals below typically 0.5Hz. From 0.5Hz and upwards, the WaveForms scope triggers correctly, however below 0.5Hz, the trigger point is unpredictable.

What can be the reason for this and is there a workaround for it?

kind regards,

Hugo

 

Share this post


Link to post
Share on other sites

2 answers to this question

Recommended Posts

  • 0
21 hours ago, attila said:

By default the Scope, with time base at 100mV/div or above, changes the run button function to scan. Change the Run mode to Repeated.
Set trigger mode to Normal, to not to enter in Auto triggered mode.

https://reference.digilentinc.com/waveforms3:refmanual#control

Setting the trigger mode to Normal solved the problem. Thanks a lot.

kind regards,

Hugo